What is an internal standard and why is it used in LC-MS/MS analyses?

Prepare for the MTLAWS Drug Testing Screening Test. Study with flashcards and multiple-choice questions, each question includes hints and explanations. Get ready to excel in your exam!

Multiple Choice

What is an internal standard and why is it used in LC-MS/MS analyses?

Explanation:
In LC-MS/MS analyses, variability from sample preparation, extraction efficiency, injection volume, and ionization (matrix effects) can cause the measured signal to wander from run to run. An internal standard is a compound that is chemically similar to the analyte of interest, added to every sample (and to calibration standards and quality controls) in a constant amount. Because it behaves similarly to the analyte during extraction, cleanup, and ionization, any fluctuations affect both compounds in a related way. By comparing the signal of the analyte to the signal of the internal standard (often using a ratio), you normalize these variations, leading to more accurate and precise quantification across samples and runs. The internal standard is typically not the analyte itself and is chosen to be distinguishable by the instrument, often a stable isotope-labeled version or a closely related compound, and it’s added in a fixed amount to ensure the ratio truly reflects changes in the analyte concentration.

In LC-MS/MS analyses, variability from sample preparation, extraction efficiency, injection volume, and ionization (matrix effects) can cause the measured signal to wander from run to run. An internal standard is a compound that is chemically similar to the analyte of interest, added to every sample (and to calibration standards and quality controls) in a constant amount. Because it behaves similarly to the analyte during extraction, cleanup, and ionization, any fluctuations affect both compounds in a related way. By comparing the signal of the analyte to the signal of the internal standard (often using a ratio), you normalize these variations, leading to more accurate and precise quantification across samples and runs. The internal standard is typically not the analyte itself and is chosen to be distinguishable by the instrument, often a stable isotope-labeled version or a closely related compound, and it’s added in a fixed amount to ensure the ratio truly reflects changes in the analyte concentration.

Subscribe

Get the latest from Examzify

You can unsubscribe at any time. Read our privacy policy